User: gar, Location: Ann Arbor, Michigan
Asked on: Wednesday, January 12, 2011 at 1:19 pm EST
Answered on: Tuesday, May 14, 2024 at 12:09 am EDT
Question: 110112-1259 ESTI need to calibrate a thermocouple somewhere around 650 deg F. I would like a material that can provide an accuracy better than +/-0.5 deg F. Anything in the range of 600 - 700 deg F is probably adequate.63-37 tin-lead solder, eutectic, is listed at 361.4 deg F. This would be better than nothing but it is a big extrapolation to 650. Also I do not know how precise the alloy content is for commericial solder.I have searched the Internet to a limited extent.One interesting reference ishttp://cmp.physics.iastate.edu/canfield/course/Calorimetry I.pdfbut has nothing to answer my specific question.Maybe weressl might have some information.When I ask this kind of question, then it leads to the question of fundamentally how do you make absolute temperature measurements? These have probably been done for many years with high accuracy, but probably not before 1850..
NIST has a temperature calibration service. Their web site is:http://ts.nist.gov/temperature/They are currently offering to calibrate sensors from -40 deg C to 1100 deg C. These services are not cheap. Their price sheet shows a fee of $2400 for calibrations at five fixed temperatures. However, their uncertainty is estimated to be 0.07 deg C ( +/- 0.13 deg F).This might be a little more accuracy than you need, but it is good to know how far you can go if necessary. Flag for review