User: DavidA, Location: Fresno, CA
Asked on: Friday, January 21, 2011 at 1:40 am EST
Answered on: Tuesday, May 14, 2024 at 12:33 am EDT
Question: I'm kinda flying blind as i have been unable to get sheets on the machines to be installed.. On that note, for a wall mounted dental X-ray unit 120volt 20amp circuit, will a 20A single pole switch be sufficient?If so, are the 'controls' for the machine typically considered to be the push button to activate the machine normally located outside of the X-ray room (and therefore the location also for the disconnectin means)? 517.72(B) 2005ed.
Circuit Protection and Disconnecting MeansYes, a 20A single pole switch would be sufficient for a wall-mounted dental X-ray unit with a 120 volt, 20-amp circuit.
Under NEC 517.72(B) (2005 edition)
- A single pole switch or circuit breaker is permitted for disconnecting the power supply to X-ray equipment.- The disconnecting means shall be located outside the X-ray room, with a clear line of sight from the operator's position.
Controls
The controls for the machine are typically considered to be the push button to activate the machine, which is usually located outside the X-ray room. This push button initiates the X-ray exposure and is part of the required safety measures to prevent accidental exposure. Flag for review