Mike Dolt Enterprises Logo Mike Dolt Enterprises

Topic: Ground Fault Simulator

User: RickLosi, Location: Connecticut

Asked on: Tuesday, July 1, 2025 at 11:22 am EDT
Answered on: Tuesday, July 1, 2025 at 11:23 am EDT

Question: I am currently getting a temperature controller with built-in GFEPD tested by Intertek to the UL 1053 Standard, I am trying to locate a Ground Fault Simulator that they can use to simulate ground leakage current at 85%, 115%, 150% and 250% at the rated Trip Setpoint of 30 mA. This is required to certify that our equipment is capable of tripping the circuit as designed. I am having difficulty locating a piece of equipment that I can buy or rent to facilitate the testing.Does anyone know of such a piece of equipment?

You're right, finding a commercial, off-the-shelf ground fault simulator with that level of precise control, especially designed for UL 1053 testing with those specific current levels and trip points, can be tricky. Most standard testers are designed for general GFCI/ALCI testing and might not offer the fine-grained control you need.Here's a breakdown of approaches and potential solutions, focusing on what you can buy/rent or build

1. Speak Directly with Intertek:
Best First Step

Before investing in anything,
directly ask the Intertek test engineer what they typically use or recommend for UL 1053 testing to this specification. They likely have a preferred method and may even have specific equipment they can use or guide you towards. This will save you time and money. They are also the most familiar with their own lab capabilities. Ask if they have a GFS.

Ask for Specific Brands/Models

Probe for actual product names and manufacturers.

2. Custom-Built Ground Fault Simulator (Likely Best Solution if off-the-shelf isn't available)



Why Custom?
The precise control you need often necessitates a custom approach. Key Components & Design



Variable AC Power Source

A stable AC power source (120V or 240V, whatever your device uses) is the foundation. A basic variac (variable transformer) can work if you don't need perfect voltage regulation.


Precision Resistor Network

This is the heart of the simulator. You need resistors that, when connected between the hot/neutral line and ground, will create the precise fault currents you need (85%, 115%, 150%, and 250% of 30mA).

Calculating Resistor Values

This is where Ohm's Law comes in (R = V/I). If your system is 120V:
30mA Trip: R = 120V / 0.03A = 4000 ohms 85% of 30mA: I = 0.0255A, R = 120V / 0.0255A = 4705.88 Ohms 115% of 30mA: I = 0.0345A, R = 120V / 0.0345A = 3478.26 Ohms 150% of 30mA: I = 0.045A, R = 120V / 0.045A = 2666.67 Ohms 250% of 30mA: I = 0.075A, R = 120V / 0.075A = 1600 Ohms

Power Rating

Calculate the power dissipated by each resistor (P = V^2 / R). Use resistors with a power rating significantly
higher than the calculated value (at least double, ideally more) to ensure they don't overheat and change resistance.

Construction

Use high-quality, precision resistors (1% tolerance or better). Wire them to a switch or rotary selector to easily select the desired fault current. Consider using a decade resistance box for greater flexibility.

Current Meter (Clamp-on Ammeter)

A clamp-on ammeter that can accurately measure mA AC current is crucial to verify the actual fault current flowing. Many multimeters have a mA range, but a clamp-on is often easier to use in this scenario. Look for one with good accuracy at low current levels.


Switching Mechanism

A robust switch (or series of switches) to apply the fault condition. This switch should be able to handle the voltage and current.

Enclosure & Safety Features

A proper enclosure to house the components and protect the user is essential. Use appropriate wiring, terminals, and safety interlocks (e.g., a fuse or circuit breaker).

Safety is paramount when working with AC voltage!
Who Can Build It? Your Internal Engineering Team

If you have electrical engineers or technicians, they are the ideal candidates to design and build this.


Contract Electronics Manufacturer (CEM)

A CEM can take your specifications and build the simulator for you.

Local Electrical Contractor (with engineering expertise)

Some electrical contractors have the skills to build custom test equipment.

3. Modified GFCI/ALCI Tester



The Challenge

Most GFCI/ALCI testers don't offer precise control over the fault current.

Possible Modification

You
might be able to modify a GFCI/ALCI tester by adding an external resistor network in series with the internal fault resistor. This would require some experimentation and careful measurement to achieve the desired current levels. This approach is risky and might damage the tester, so proceed with caution and only if you have significant electrical engineering expertise. I highly recommend against this.

4. Potential Rental Options



Test Equipment Rental Companies

Companies like United Rentals or Electro Rent might have equipment that can be adapted, but you'll need to call and speak to their technical experts. Be prepared to explain your specific needs and the UL 1053 requirements. It's highly unlikely they will have such specific equipment, but it is worth a try.

Important Considerations



Calibration

Any simulator you build or modify
must be accurately calibrated. Use a calibrated current meter to verify the fault current at each setting.

Safety

Working with AC voltage is dangerous. Ensure the simulator is properly insulated, grounded, and fused. Wear appropriate safety gear. If you aren't comfortable working with AC voltage, hire a qualified electrician or engineer.

Documentation

Keep thorough documentation of the simulator's design, construction, and calibration. This documentation will be important for the testing lab.


UL 1053 Specifics

UL 1053 might have specific requirements for the test setup. Refer to the standard itself or ask the Intertek engineer.*

Don't Overthink It
: Simulating the fault current does not need to be as complex as you might think. The simulator only needs to source the current to the controller to perform the test as required by the standard.

In summary, the best approach is to


Talk to Intertek first!
They are your primary resource.2. Consider building a custom simulator using precision resistors and a current meter. This is the most likely path to success.3. Explore rental options as a secondary approach.4. Avoid modifying existing GFCI/ALCI testers unless you have significant electrical engineering expertise.

Good luck! Flag for review

Previous Question View on the official Mike Holt Forum Next Question